Romer Arm Hardprobe to Scanner Alignment

Started by Jarhead, February 14, 2013, 06:53:59 AM

Previous topic - Next topic

Jarhead

We have a Romer Infinite Arm with the Scanshark scanner.  How can i align the two sensors to one another within the software?  Sometimes we try use the probe to align to the part befoe scanning, but i am seeing an offset between the sensors.  I belive their should be a way within the software (Cimcore utilites) to zero one to the other.

TIA.
-James

Semper Fi

amoreno2000

Depending of the version of RDS you have,  It will guide you on how to calibrate the scanner using a reference sphere.   Normally you use the master probe to measure the reference sphere (Like 1" dia) and just follow the steps.

JM