Title: Romer Arm Hardprobe to Scanner Alignment
Post by: Jarhead on February 14, 2013, 06:53:59 AM
We have a Romer Infinite Arm with the Scanshark scanner.  How can i align the two sensors to one another within the software?  Sometimes we try use the probe to align to the part befoe scanning, but i am seeing an offset between the sensors.  I belive their should be a way within the software (Cimcore utilites) to zero one to the other.

TIA.
Title: Re: Romer Arm Hardprobe to Scanner Alignment
Post by: amoreno2000 on May 26, 2015, 09:54:27 PM
Depending of the version of RDS you have,  It will guide you on how to calibrate the scanner using a reference sphere.   Normally you use the master probe to measure the reference sphere (Like 1" dia) and just follow the steps.

JM